Version 0.5tc2 Touchscreen Calibration

Posted: Thu Jan 24, 2019 12:17 pm
by Logan
.

In version 0.5tc2 there is new touchscreen calibration file

When you upload new software, you will notice calibration squares that appear in the 4 corners of the touch screen



If you install the new version, and are experiencing difficulties, try the following:
  1. Restart Tester by holding Left and Right button together for roughly 10 seconds
  2. Then try to install new software following our Software Update Guide
  3. After installation is complete, you should see the calibration screen with red numbers in each square
  4. Touch each square in order and they will turn green - after all 4 corners are set, the display will be calibrated

If for some reason you do not see a calibration screen after installing a new version, then follow these steps:
  1. Connect tester to computer via USB cable
  2. When the main menu appears, press up and down button together
  3. In computer directory you will see a new USB drive appear
  4. In that USB drive you will delete all files named “calibrated“ (you need to see hidden files - in windows folder options, select the checkbox titled “Show hidden files, folders and drives”)
  5. After restarting the tester, you should see the calibration screen and can follow the steps above.